T500 Temperature probe issue
Posted:
Tue Feb 02, 2021 9:52 pm
by Stillworkin
I did a calibration check on my T500 temp probe using an accurate mercury thermometer and discovered that at 85 deg C, the T500 probe over reads by 5 deg C.
The temp difference at 50 deg C is zero!
Does anyone know how to re-calibrate the T500 probe or just chill out?
TIA
Re: T500 Temperature probe issue
Posted:
Wed Feb 03, 2021 8:44 am
by wynnum1
How old is the probe because that looks to be very inaccurate but if measures constant wrong temperature could just adjust for the difference.
Re: T500 Temperature probe issue
Posted:
Wed Feb 03, 2021 4:46 pm
by Stillworkin
wynnum1 wrote:How old is the probe because that looks to be very inaccurate but if measures constant wrong temperature could just adjust for the difference.
Thanks wynnum1, I think the probe is about 5 yrs old...I think I'll just adjust for the error. :handgestures-thumbupleft:
Re: T500 Temperature probe issue
Posted:
Fri Feb 12, 2021 11:00 am
by B-Man
when I ran a T500 the temp probe only showed me it was running stable.
I took no notice of the actual temp it varied so much and from what everyone else reckons "they run great at xx temp" mine ran great at 38°c but others said i was lying because they couldn't get a stream at that and they had to run at 60°c.
Just run based on the stream output and you wont have an issue
the 4" bubbler I have now doesn't even have a temp probe. I was going to put one in while building it but just wanted it running so haven't finished a few things off and the more I have used it the more I have decided its not needed.